Which combination of parameters will create the highest heating?

Prepare for the MRSE exam with comprehensive study tools including flashcards and multiple-choice questions. Each question provides explanations and hints. Enhance your readiness and confidence for the test!

The correct choice indicates a combination of parameters that maximizes the potential for RF (radiofrequency) heating in an MRI environment. In MR imaging, several factors contribute to RF heating, including the repetition time (TR), pulse duration, and flip angle.

A shorter TR allows for more frequent pulses of radiofrequency energy to be delivered to the tissue, leading to increased energy deposition and heating. The flip angle also significantly impacts heating; a higher flip angle means greater excitation of the spins in the tissue, which increases the energy transferred to the tissue. Additionally, longer pulse durations can result in a longer exposure to RF energy, further increasing the potential for heating.

Option D, with a TR of 1500 ms, a pulse duration of 2 ms, and a flip angle of 180 degrees, presents a scenario with a relatively short repetition time and a high flip angle. This combination allows for a greater amount of RF energy being applied to the tissue with more substantial excitation effects, thus enhancing the heating effect.

In contrast, the other options feature either longer TRs or lower flip angles that would not achieve the same level of energy deposition as the parameters in the correct choice. By selecting parameters that yield more effective RF energy absorption and excitation, option D effectively